USRP Hardware Driver and USRP Manual  Version: 4.6.0.0-7-gece7c4811
R&D Testing Procedures
All defined R&D test procedures are listed here. These tests are meant as a tool for Ettus R&D to enable faster and more reliable development. Note these tests are no replacement for manufacturing or production tests, and should not be treated as such. Instead, they are meant to catch common failure modes during development. As a result, test definitions are fairly light-weight.
GPSDO Tests
Test Code Device Peripherals Manual Test Procedure Automatic Test Procedure
GPS-X310-TCXO-v1 USRP X310 Jackson Labs TCXO GPSDO: Manual Test Procedure GPSDO: Automatic Test Procedure
GPS-X310-OCXO-v1 USRP X310 Jackson Labs OCXO GPSDO: Manual Test Procedure GPSDO: Automatic Test Procedure
GPS-X300-TCXO-v1 USRP X300 Jackson Labs TCXO GPSDO: Manual Test Procedure GPSDO: Automatic Test Procedure
GPS-X300-OCXO-v1 USRP X300 Jackson Labs OCXO GPSDO: Manual Test Procedure GPSDO: Automatic Test Procedure
GPS-B200-TCXO-v1 USRP B200 Jackson Labs TCXO GPSDO: Manual Test Procedure GPSDO: Automatic Test Procedure
GPS-B210-TCXO-v1 USRP B210 Jackson Labs TCXO GPSDO: Manual Test Procedure GPSDO: Automatic Test Procedure
Recommendations
For cursory testing, not all tests within a device family are required (e.g., only testing the OCXO on any X-Series, and testing the TCXO on any B-Series is sufficient).
The following tests are recommended for a minimum test (N stands for the latest version of this test):
  • One of GPS-X310-OCXO-vN or GPS-X300-OCXO-vN
  • One of GPS-B210-TCXO-vN or GPS-B200-TCXO-vN
Requirements
All of these tests require a device that is GPSDO capable (e.g., X3x0, B2x0, N2x0). For those devices that have a separate GPS component (such as the Jackson Labs GPSDOs), this component is also required (called the "peripheral" in the following).
  1. Without connecting the peripheral to the device, run uhd_usrp_probe on the device and verify that the lack of GPSDO is correctly reported under "sensors". No warning or error must be printed.
  2. This and the following tests are run with the peripheral connected: Run uhd_usrp_probe and verify that the GPSDO is correctly reported under "sensors". Power down the device before connecting the peripheral. The GPSDO must be reported found, and no error or warning must be printed.
  3. OCXO only: Without connecting the GPS antenna input, run utils/query_gpsdo_sensors. To pass, it must report the GPSDO as found, lock to the external reference, but then report not being locked to GPS. The tool will report a valid GPS time, and a string such as "GPS and UHD Device time are aligned" in case of success.
  4. Connect a GPS antenna to the input and make sure it is in a position to receive GPS satellite data. Confirm that GPS lock is reported by running utils/query_gpsdo_sensors within 20 minutes of connecting the antenna. The tool query_gpsdo_sensors will print a string such as "GPS Locked" in case of success.
All of these tests must pass for a 'pass' validation.
tbd
Devtests
DEVTEST-X310-XG-v1 USRP X310 None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-X310-HG-v1 USRP X310 None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-X300-XG-v1 USRP X300 None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-X300-HG-v1 USRP X300 None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-E310-SG1-v1 USRP E310-SG1 None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-E310-SG3-v1 USRP E310-SG3 None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-B200-v1 USRP B200 None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-B210-v1 USRP B210 None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-B200m-v1 USRP B200mini None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-B205m-v1 USRP B205mini None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-N310-v1 USRP N310 None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
DEVTEST-E320-v1 USRP E320 None Devtest: Manual Test Procedure Devtest: Automatic Test Procedure
The devtests are hardware tests built in to the UHD make system. They can be run directly from the build directory and require no configuration. Devtests are designed to always run, regardless of the actual device configuration. This means, by definition, that devtests cannot require special cabling, specific daughtercards, etc.
Note: The actual devtests can change, since they're part of the code. This does not require a version bump on the test code.
Requirements
Devtests are only defined for some devices. When running a devtest, all peripherals must be disconnected (e.g., no daughterboards on the X-Series, no GPSDOs on the B- and X-Series).
Running these tests requires the yaml package. On Ubuntu, run sudo apt-get install python-yaml to install the Python 2 version of the YAML library.
X3x0 procedure
  1. Make sure no peripherals are connected to the device (no daughterboards, no GSPDO, front panel GPIO is unconnected).
  2. When testing the HG image, run a test once for each connection (1 GigE and 10 GigE). When testing the XG image, a test on either connection (SFP0 or SFP1) is sufficient. In both cases, also test via PCIe.
  3. When the device is connected, simply run make test_x3x0 from the command line in the build directory. Multiple devices connected will all get tested, there is no requirement to only connect a single device at a time (because devtest will run sequentially anyway).
  4. Devtest must report no failures for a 'pass' validation.
B2xx procedure
Note: The test codes with an 'm' suffix refer to B200mini and B205mini, respectively.
  1. Make sure no peripherals are connected to the device (no GPSDO if applicable, GPIO pins unconnected)
  2. Test once via USB3, once via USB2.
  3. Simply run make test_b2xx
E310 procedure
  1. Make sure GPIO pins are unconnected.
  2. Tests need to be run natively on the device. If the build environment is available on the device, running make test_e3xx is sufficient.
  3. In general, there is no build environment on the device (e.g. when doing a typical sshfs mount of an environment). In this case, copy the contents of the devtest directory onto the device, and run the following command (the environment variables need to point to the location of the devtest code, the location of the UHD build, and where you want to store log files, respectively):
    $DEVTEST_DIR/run_testsuite.py --src-dir $DEVTEST_DIR \
                                  --devtest-pattern e3xx \
                                  --build-type na \
                                  --build-dir $BUILD_DIR \
                                  --args type=e3x0 \
                                  --log-dir $LOG_DIR
E320 procedure
  1. Make sure no peripherals are connected to the device (no GSPDO, front panel GPIO is unconnected).
  2. Run tests for both 1G and XG image.
  3. When the device is connected, simply run make test_e320 from the command line in the build directory.
As all these tests can be run unsupervised, they can be run automatically given the correct device setup. The return code of the tests can be used to check for pass/fail conditions (return code 0 means 'pass').
FPGA: Testing through Simulations
Test benches provide a faster way to verify the design through simulations.
FPGATB-v1 None None Manual Test Procedure Automatic Test Procedure
Requirements
These tests are simulations and do not need any device. Vivado 15.4 should be installed.
  1. Go to the fpga directory depending on which test needs to be run.
    1. NoC block test benches: Most of the NoC blocks have a test bench written in System Verilog that provides stimuli to the noc_block to verify it. The test bench for a block resides in <fpga-dir>/usrp3/lib/rfnoc/‍*_tb.
    2. Running unit test benches: A few sub-blocks like noc-shell and sine_tone are used within the bigger noc_blocks. They have their own test benches. Their test benches reside in <fpga-dir>/usrp3/lib/sim/rfnoc/‍*.
    3. Radio test bench: The radio test bench resides in <fpga-dir>/usrp3/lib/radio/noc_block_radio_core_tb/.
    4. Device specific test benches: IPs specific to a device have test benches that exist in <fpga-dir>/usrp3/top/x300/sim/\*. e.g. DMA testbench, PCIe, etc.
  2. Setup the environment by running source <fpga-dir>/usrp3/top/<device>/setupenv.sh.
  3. In the test bench directory and run the test bench by 'make xsim' or 'make vsim'.
  4. All of these tests must report no failure with a 'PASS' validation. Example testbench output:
TESTBENCH STARTED: noc_block_skeleton
[TEST CASE 1] (t=000000000) BEGIN: Wait for Reset...
[TEST CASE 1] (t=000001002) DONE... Passed
[TEST CASE 2] (t=000001002) BEGIN: Check NoC ID...
Read Skeleton NOC ID: 1234000000000000
[TEST CASE 2] (t=000001238) DONE... Passed
[TEST CASE 3] (t=000001238) BEGIN: Connect RFNoC blocks...
Connecting noc_block_tb (SID: 1:0) to noc_block_skeleton (SID: 0:0)
Connecting noc_block_skeleton (SID: 0:0) to noc_block_tb (SID: 1:0)
[TEST CASE 3] (t=000005457) DONE... Passed
[TEST CASE 4] (t=000005457) BEGIN: Write / readback user registers...
[TEST CASE 4] (t=000006888) DONE... Passed
[TEST CASE 5] (t=000006888) BEGIN: Test sequence...
[TEST CASE 5] (t=000007403) DONE... Passed
TESTBENCH FINISHED: noc_block_skeleton
- Time elapsed: 7500 ns
- Tests Expected: 5
- Tests Run: 5
- Tests Passed: 5
Result: PASSED
Failing tests can be debugged by checking the waveform in a Vivado GUI by running 'make GUI=1 xsim'. More details on debugging: https://kb.ettus.com/Debugging_FPGA_images
Go to <fpga-dir>/usrp3/ and run 'build.py xsim all'. All tests should report 'PASS'.
FPGA DSP Verification
FPGADSPVERIF-X310-HG-v1 USRP X310 2x UBX FPGA DSP Verification: Manual Test Procedure FPGA DSP Verification: Automatic Test Procedure
FPGADSPVERIF-X310-XG-v1 USRP X310 2x UBX FPGA DSP Verification: Manual Test Procedure FPGA DSP Verification: Automatic Test Procedure
FPGADSPVERIF-X300-HG-v1 USRP X300 2x UBX FPGA DSP Verification: Manual Test Procedure FPGA DSP Verification: Automatic Test Procedure
FPGADSPVERIF-X300-XG-v1 USRP X300 2x UBX FPGA DSP Verification: Manual Test Procedure FPGA DSP Verification: Automatic Test Procedure
FPGADSPVERIF-E310-SG1-v1 USRP E310 SG1 None FPGA DSP Verification: Manual Test Procedure FPGA DSP Verification: Automatic Test Procedure
FPGADSPVERIF-E310-SG3-v1 USRP E310 SG3 None FPGA DSP Verification: Manual Test Procedure FPGA DSP Verification: Automatic Test Procedure
FPGADSPVERIF-E320-v1 USRP E320 None FPGA DSP Verification: Manual Test Procedure FPGA DSP Verification: Automatic Test Procedure
Requirements
  • Signal generator and spectrum analyzer
  • X300 & X310 with 2x UBX daughterboard
  • E310 SG1 & SG3 with SSH access
  • E320 with SSH access
This procedure tests the DDC and DUC signal quality and the block's capability to change sample rate while streaming.
RX testing
  1. Run calibration on device, if applicable
  2. Using a signal generator, inject a sine tone into RX channel 0 at 915.5 MHz @ -40 dBm
  3. Inspect the received spectrum using uhd_fft
    • X3x0: uhd_fft -f 915e6 -s 10e6 -g 10
    • E31x: uhd_fft -f 915e6 -s 2e6 -g 50
    • E320: uhd_fft -f 915e6 -s 15.36e6 -g 50
    • Embedded devices will require either using network mode or using X forwarding over ssh to run the app natively
  4. In the GUI, inspect the spectrum. There should be a strong tone at the test tone frequency. There may be a small tone at the carrier frequency due to DC offset and a quadrature image due to IQ imbalance.
  5. Check the input tone frequencies outlined below. The tone should shift from left to right as the frequency changes and may have some amplitude variation, especially at the band edges.
    • X3x0: 910 MHz to 920 MHz in 1 MHz steps
    • E3xx: 914 MHz to 916 MHz in 200 kHz steps
  6. Set input tone back to 915.5 MHz. Check the sampling rate as outlined below. The spectrum should reflect the change in sample rate.
    • X3x0: 1, 5, 20, 33.333, 50, 66.666, 100, 200 MHz
    • E3x0: 0.1, 0.5, 1, 1.143, 1.684 MHz
  7. Repeat on each RX channel of the device.
  8. This test fails if:
    • DC offset and IQ imbalance tones are unusually large
    • There are any other strong tones or spectrum distortion
    • The spectrum changes significantly between frequencies or sample rates
      • An initial transient distortion is acceptable
      • Amplitude variation on the order of +/-10 dB is acceptable
    • Console reports any of the following:
      • Overruns 'O' if continuous and not due to host computer's lack of processing performance
      • Dropped packets 'D'
      • Sequence number errors 'S'
      • Timeouts
TX testing
  2. Using uhd_siggen_gui, generate a sine tone TX channel 0 at 915.5 MHz:
    • X3x0: uhd_siggen_gui -f 915e6 -s 10e6 -g 10 -x 500e3 --sine
    • E31x: uhd_siggen_gui -f 915e6 -s 2e6 -g 50 -x 500e3 --sine
    • E320: uhd_siggen_gui -f 915e6 -s 15.36e6 -g 50 -x 500e3 --sine
  3. Using a spectrum analyzer, inspect the output spectrum. There should be a strong tone at the test tone frequency. There may be a small tone at the carrier frequency due to DC offset and a quadrature image due to IQ imbalance.
  4. Using the GUI, test the follow offset frequencies. The tone should shift from left to right as the frequency changes and may have some amplitude variation, especially at the band edges.
    • X3x0: -5 to +5 MHz in 1 MHz steps
    • E3xx: -1 to +1 MHz in 200 kHz steps
  5. Set output tone offset back to 500e3. Change sampling rate as outlined below. The spectrum should not significantly differ between sample rates.
    • E3xx: 0.1, 0.5, 1, 1.143, 1.684 MHz
  6. Repeat on each TX channel of the device
  7. This test fails if:
    • The spectrum changes significantly between sample rates
      • Underruns 'U' if continuous and not due to host computer's lack of processing performance
      • Late packets 'L'
tbd
FPGA Functional Verification
FPGAFUNCVERIF-X310-HG-v1 USRP X310 2x UBX FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-X310-XG-v1 USRP X300 2x UBX FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-X300-HG-v1 USRP X310 2x UBX FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-X300-XG-v1 USRP X300 2x UBX FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-E310-SG1-v1 USRP E310 SG1 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-E310-SG3-v1 USRP E310 SG3 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-N320-HG-v1 USRP N320 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-N320-XG-v1 USRP N320 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-N310-HG-v1 USRP N310 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-N310-XG-v1 USRP N310 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-N300-HG-v1 USRP N300 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-N300-XG-v1 USRP N300 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-E320-1G-v1 USRP E320 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-E320-XG-v1 USRP E320 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
FPGAFUNCVERIF-E320-device-v1 USRP E320 None FPGA Functional Verification: Manual Test Procedure FPGA Functional Verification: Automatic Test Procedure
The FPGA functional verification tests exercise the Digital Downconverter (DDC), Digital Upconverter (DUC), and Radio Core RFNoC blocks.
Requirements
  • X300 & X310 with two daughterboards
    • 2x UBX recommended
    • HG tests require a single 10 GigE connection, XG requires two for the 2x RX 200 MSPS test
    • 1 GigE and PCIe adapters and cabling for optional tests
  • E310: SG1 & SG3 with SSH access
  • N310: No special requirements
  • E320: SFP connection to run network mode
This procedure verifies that the DDC, DUC, and Radio Core can run at various sample rates and channel configurations without any data flow issues.
  1. Run benchmark_rate using the parameters outlined in the tables below
  2. Unless otherwise noted, to pass each test:
    • Benchmark rate must run without reporting any of the following:
      • Underruns 'U'
      • Overruns 'O'
      • Late commands 'L'
      • Timeouts
    • Appropriate TX/RX LEDs must be illuminated
  3. Unless specified in 'Notes' column, use default values for unlisted parameters
  4. Example commands:
    • X3x0: benchmark_rate --tx_rate 1e6 --rx_rate 1e6 --channels 0,1 --duration 120
    • E3xx: benchmark_rate --args="master_clock_rate=10e6" --tx_rate 1e6 --rx_rate 1e6 --channels 0,1 --duration 120
USRP X3x0: 10 GigE Interface
  • Required images to test: X310 HG
  • Optional images to test: X310 XG, X300 HG, X300 XG
  • Note: On TX tests, initial Us within the first 5 seconds can be ignored and do not fail the test
Channels Sample Rates Duration Notes
1x RX 10e6, 50e6, 100e6, 200e6 60 Test both channels
2x RX 10e6, 50e6, 100e6 60
2x RX 200e6 60 2x 10G, XG only
1x TX 10e6, 50e6, 100e6, 200e6 60 Test both channels
2x TX 10e6, 50e6, 100e6 60
1x RX & 1x TX 10e6, 50e6, 100e6 60 Test both channels
1x RX & 1x TX 200e6 60 Use channel 0
2x RX & 2x TX 10e6, 50e6 60
1x RX & 1x TX 200e6 600 Use channel 1
2x RX & 2x TX 100e6 600
USRP X3x0: 1 GigE Interface
  • Required images to test: None
  • Optional images to test: X310 HG, X310 XG, X300 HG, X300 XG
1x RX 1e6, 10e6, 25e6 60
2x RX 1e6, 10e6 60
1x TX 1e6, 10e6, 25e6 60
2x TX 1e6, 10e6 60
1x RX & 1x TX 1e6, 10e6, 25e6 60
2x RX & 2x TX 1e6, 10e6 60
USRP X3x0: PCIe Interface
USRP E3xx Device (Embedded Mode)
  • E320 and E310 SG3 are required tests, E310 SG1 is optional
Channels Master Clock Rate Sample Rate Duration Notes
1x RX 10e6 1e6 60 Test both channels
1x RX 61.44e6 3.84e6 60 Test both channels
1x TX 10e6 1e6 60 Test both channels
1x TX 61.44e6 3.84e6 60 Test both channels
2x RX 10e6 1e6 60
2x RX 30.72e6 1.92e6 60
2x TX 10e6 1e6 60
2x TX 30.72e6 1.92e6 60
1x RX & 1x TX 10e6 1e6 60 Test both channels
1x RX & 1x TX 30.72e6 3.84e6 60 Use channel 1
2x RX & 2x TX 10e6 1e6 60
2x RX & 2x TX 30.72e6 1.92e6 60
1x RX & 1x TX 30.72e6 3.84e6 600 Use channel 0
2x RX & 2x TX 30.72e6 1.92e6 600
USRP E320: 1 GigE Interface
1x RX 15.36e6 15.36e6 60 Test both channels
1x TX 15.36e6 15.36e6 60 Test both channels
2x RX 61.44e6 7.68e6 60
2x TX 61.44e6 7.68e6 60
1x RX & 1x TX 30.72e6 7.68e6 60 Test both channels
1x RX & 1x TX 61.44e6 1.92e6 60 Use channel 1
2x RX & 2x TX 30.72e6 3.84e6 60
USRP E320: 10 GigE Interface
1x RX 61.44e6 1.92e6, 61.44e6 60 Test both channels
1x TX 61.44e6 1.92e6, 61.44e6 60 Test both channels
2x RX 30.72e6 1.92e6, 30.72e6 60
2x TX 30.72e6 1.92e6, 30.72e6 60
1x RX & 1x TX 61.44e6 1.92e6, 30.72e6 60 Test both channels
2x RX & 2x TX 30.72e6 1.92e6, 30.72e6 60
1x RX & 1x TX 61.44e6 1.92e6, 30.72e6 600 Use channel 0
2x RX & 2x TX 30.72e6 1.92e6, 30.72e6 600
USRP N300/N310: 1 GigE Interface
  • Required images to test: HG
Channels Master Clock Rate Sample Rates Duration Notes
1x RX 125e6 1.25e6 60 One test each for all 4 channels
1x RX 122.88e6 1.2288e6 60 One test each for all 4 channels
1x RX 153.6e6 1.536e6 60 One test each for all 4 channels
1x TX 125e6 1.25e6 60 One test each for all 4 channels
1x TX 122.88e6 1.2288e6 60 One test each for all 4 channels
1x TX 153.6e6 1.536e6 60 One test each for all 4 channels
2/3/4x RX 125e6 1.25e6 60 3 tests total
2/3/4x RX 122.88e6 1.2288e6 60 3 tests total
2/3/4x RX 153.6e6 1.536e6 60 3 tests total
2/3/4x TX 125e6 1.25e6 60 3 tests total
2/3/4x TX 122.88e6 1.2288e6 60 3 tests total
2/3/4x TX 153.6e6 1.536e6 60 3 tests total
4x RX & 4x TX 125e6 1.25e6 60 Drop to 2 channels for N300
4x RX & 4x TX 122.88e6 1.2288e6 60 Drop to 2 channels for N300
4x RX & 4x TX 153e6 1.536e6 60 Drop to 2 channels for N300
USRP N300/N310: 10 GigE Interface
  • Required images to test: N310 HG + XG
1x RX 125e6 1.25e6, 125e6 60 One test each for all 4 channels
1x RX 122.88e6 1.2288e6, 122.88e6 60 One test each for all 4 channels
1x RX 153.6e6 1.536e6, 153.6e6 60 One test each for all 4 channels
1x TX 125e6 1.25e6, 125e6 60 One test each for all 4 channels
1x TX 122.88e6 1.2288e6, 122.88e6 60 One test each for all 4 channels
1x TX 153.6e6 1.536e6, 153.6e6 60 One test each for all 4 channels
2x RX 125e6 1.25e6, 125e6 60
2x RX 122.88e6 1.2288e6, 122.88e6 60
2x RX 153.6e6 1.536e6 60
3x RX 125e6 1.25e6 60 N310 only
3x RX 122.88e6 1.2288e6 60 N310 only
3x RX 153.6e6 1.536e6 60 N310 only
2x TX 125e6 1.25e6, 12.5e6 60
2x TX 122.88e6 1.2288e6, 12.288e6 60
2x TX 153.6e6 1.536e6, 15.36e6 60
3x TX 125e6 1.25e6 60 N310 only
3x TX 122.88e6 1.2288e6 60 N310 only
3x TX 153.6e6 1.536e6 60 N310 only
4x RX 125e6 1.25e6, 62.5e6 60 N310 only
4x TX 125e6 1.25e6, 12.5e6 60 N310 only
4x RX & 4x TX 125e6 1.25e6, 62.5e6 60 Drop to 2 channels for N300
4x RX & 4x TX 122.88e6 1.2288e6, 61.44e6 60 Drop to 2 channels for N300
4x RX & 4x TX 153e6 1.536e6, 76.8e6 60 Drop to 2 channels for N300
4x RX & 4x TX 125e6 62.5e6 600 Drop to 2 channels for N300
4x RX & 4x TX 122.88e6 61.44e6 600 Drop to 2 channels for N300
4x RX & 4x TX 153e6 76.8e6 600 Drop to 2 channels for N300
4x RX & 4x TX 125e6 125e6 RX, 62.5e6 TX 60 Use dual 10GigE, N310 XG only
4x RX & 4x TX 122.88e6 122.88e6 RX, 61.44e6 TX 60 Use dual 10GigE, N310 XG only
2x RX & 2x TX 125e6 125e6 RX, 62.5e6 TX 60 Use dual 10GigE, N300 XG only
2x RX & 2x TX 122.88e6 122.88e6 RX, 61.44e6 TX 60 Use dual 10GigE, N300 XG only
2x RX & 2x TX 153e6 153e6 RX, 76.8e6 TX 60 Use dual 10GigE, N300 XG only
USRP N320/N321: 1 GigE Interface
  • Note: The following tests need to be completed only on either N320 OR N321
1x RX 250e6 2.5e6 60 One test each for both channels
1x RX 245.76e6 2.4576e6 60 One test each for both channels
1x RX 200e6 2e6 60 One test each for both channels
1x TX 250e6 2.5e6 60 One test each for both channels
1x TX 245.76e6 2.4576e6 60 One test each for both channels
1x TX 200e6 2e6 60 One test each for both channels
2x RX 250e6 2.5e6 60
2x RX 245.76e6 2.4576e6 60
2x RX 200e6 2e6 60
2x TX 250e6 2.5e6 60
2x TX 245.76e6 2.4576e6 60
2x TX 200e6 2e6 60
2x RX & 2x TX 250e6 2.5e6 60
2x RX & 2x TX 245.76e6 2.4576e6 60
2x RX & 2x TX 200e6 2e6 60
USRP N320/N321: 10 GigE Interface
  • Required images to test: N320 HG + XG
1x RX 250e6 2.5e6 , 125e6 60 One test each for both channels
1x RX 245.76e6 2.4576e6, 122.88e6 60 One test each for both channels
1x RX 200e6 2e6 , 200e6 60 One test each for both channels
1x TX 250e6 2.5e6 , 125e6 60 One test each for both channels
1x TX 245.76e6 2.4576e6 , 122.88e6 60 One test each for both channels
1x TX 200e6 2e6 , 100e6 60 One test each for both channels
2x RX 250e6 2.5e6 , 125e6 60
2x RX 245.76e6 2.4576e6, 122.88e6 60
2x RX 200e6 2e6 , 100e6 60
2x TX 250e6 62.5e6 60
2x TX 245.76e6 61.44e6 60
2x TX 200e6 100e6 60
2x RX & 2x TX 250e6 125e6 RX, 62.5e6 TX 600
2x RX & 2x TX 245.76e6 122.88e6 RX, 61.44e6 TX 600
2x RX & 2x TX 200e6 100e6 RX, 66.67e6 TX 600
2x RX & 2x TX 250e6 125e6 RX, 83.33e6 TX 600 Use dual 10GigE, N320/1 XG only
2x RX & 2x TX 245.76e6 122.88e6 RX, 81.92e6 TX 600 Use dual 10GigE, N320/1 XG only
2x RX & 2x TX 200e6 200e6 RX, 100e6 TX 600 Use dual 10GigE, N320/1 XG only
2x RX & 2x TX 250e6 250e6 600 Dual 10GigE, N320/1 XG, DPDK only
2x RX & 2x TX 245.76e6 245.76e6 600 Dual 10GigE, N320/1 XG, DPDK only
2x RX & 2x TX 200e6 200e6 600 Dual 10GigE, N320/1 XG, DPDK only
In all cases, make sure UHD is compiled in 'Release' mode (highest optimization), and that all NIC and kernel are set to optimal (CPU governor, ring buffer settings, ...).
X310/X300
The X310/X300 tests depend on the FPGA image to be tested.
HG
-Connect a 1GigE cable into port 0 and a 10GigE cable into port 1. -The following command must pass:
      $ usrp_fpga_funcverif x3x0hg -a 192.168.40.2 -2 192.168.10.2 -p /path/to/examples
XG
-Connect 10GigE cables to both ethernet ports. -The following command must pass:
      $ usrp_fpga_funcverif x3x0xg -a 192.168.40.2 -2 192.168.30.2 -p /path/to/examples
N310/N300
The N310/N300 tests depend slightly on the type of FPGA image to be tested. All calls to usrp_fpga_funcverif.py need to be adapted to ensure the correct IP addresses and paths to the examples. Also, replace n310 with n300 where appropriate.
HG
  • Connect a 1GigE cable on SFP0, and a 10 GigE cable on SFP1.
  • The following command must pass:
       $ usrp_fpga_funcverif n310hg -a 192.168.20.2 -2 192.168.10.2 -p /path/to/examples
XG
  • Connect a 10GigE cable on both SFP0 and SFP1.
       $ usrp_fpga_funcverif n310xg -a 192.168.10.2 -2 192.168.20.2 -p /path/to/examples
HA
       $ usrp_fpga_funcverif n310ha -2 192.168.10.2 -p /path/to/examples
XA
  • Connect a 10GigE cable on SFP0
       $ usrp_fpga_funcverif n310xa -a 192.168.10.2 -p /path/to/examples
WX
  • Connect a 10GigE cable on SFP1
       $ usrp_fpga_funcverif n310wx -a 192.168.20.2 -p /path/to/examples
N320/N321
The N320/N321 tests depend slightly on the type of FPGA image to be tested. All calls to usrp_fpga_funcverif.py need to be adapted to ensure the correct IP addresses and paths to the examples. Also, the following test need to be run only on either N320 OR N321.
HG
       $ usrp_fpga_funcverif n320hg -a 192.168.20.2 -2 192.168.10.2 -p /path/to/examples
XG
       $ usrp_fpga_funcverif n320xg -a 192.168.10.2 -2 192.168.20.2 -p /path/to/examples
XQ
  • Connect a QSFP+ cable, carrying 2x10GigE, to the QSFP+ port
       $ usrp_fpga_funcverif n320xq -a 192.168.10.2 -2 192.168.20.2 -p /path/to/examples
AQ
       $ usrp_fpga_funcverif n320aq -a 192.168.10.2 -2 192.168.20.2 -p /path/to/examples
WX
  • Connect a 10GigE cable on SFP1.
       $ usrp_fpga_funcverif n320wx -a 192.168.20.2 -p /path/to/examples
E310
The E310 tests need to be run on the device in the embedded mode.
Embedded mode
  • Login into the device.
       $ usrp_fpga_funcverif e3xxdev -a 127.0.0.1 -p /path/to/examples
E320
The E320 tests depend on the FPGA image to be tested.
1G
  • Connect a 1GigE cable on the SFP port.
       $ usrp_fpga_funcverif e3201g -a 192.168.10.2 -p /path/to/examples
XG
  • Connect a 10GigE cable on the SFP port.
       $ usrp_fpga_funcverif e320xg -a 192.168.10.2 -p /path/to/examples
Embedded mode
Phase alignment tests
PHASE-Twin-RX-v1 2xTwinRX 1xX3x0 + LOSharing cables X3x0 with TwinRX Automatic phase alignment testing (Receiver)
PHASE-UBX-40-RX-v1 2xUBX-40 2xX3x0 X3x0 with SBX or UBX Automatic phase alignment testing (Receiver)
PHASE-UBX-160-RX-v1 2xUBX-160 2xX3x0 X3x0 with SBX or UBX Automatic phase alignment testing (Receiver)
PHASE-SBX-40-RX-v1 2xSBX-40 2xX3x0 X3x0 with SBX or UBX Automatic phase alignment testing (Receiver)
PHASE-SBX-120-RX-v1 2xSBX-120 2xX3x0 X3x0 with SBX or UBX Automatic phase alignment testing (Receiver)
PHASE-N2x0-MIMO-v1 2x N2x0 + MIMO cable 2x SBX N2x0 MIMO with SBX Automatic phase alignment testing (Receiver)
Device Frequency Range Number of bands
TwinRX 10 - 6000 MHz 12
UBX-{160, 40} 10 - 6000 MHz 12
SBX-{120, 40} 400 - 4400 MHz 7
Phase alignment testing is necessary to verify device synchronization across multiple daughter- and motherboards is working as expected for CBX, SBX and UBX daughterboards. To enable efficient Phase alignment testing a python based test script exists tools/usrptest. It is required for testing RX testcases and later may be required to perform TX testcases.
To test phase alignment we measure phase offset between DUTs at an offset of 2 MHz offset from the selected center frequency. The phase difference for a given center frequency has to stay the same across retunes and power cycles of the DUT.
Correct synchronization with PPS and 10 MHz references is required for these tests.
Manual phase alignment testing (Receiver)
Equipment Required
  • Octoclock-G
  • Signal Generator
  • 2-way splitter that covers frequency range for daughterboard (4-way for TwinRX)
  • 5+ SMA Cables
Software Required
  • UHD
X3x0 with TwinRX
  1. Make sure correct FPGA image is loaded on X3x0.
  2. Place first daughterboard in slot A and second daughterboard in slot B.
  3. Connect LO sharing cables between boards.
  4. Connect host to device via 1 GbE, 10 GbE, or PCIe.
  5. Connect 10 MHz and PPS from Octoclock-G to X3x0.
  6. Connect Signal Generator to input of 4-way splitter and outputs of the splitter to the 2 RX ports on each daughterboard.
  7. Set Signal Generator output power at -30 dBm.
  8. From the top of the UHD source, run the command:
    • ./tools/usrptest/uhd_phase_alignment.py --spec "A:0 A:1 B:0 B:1" --channels 0,1,2,3 --sync pps --time-source external --clock-source external -s 5e6 -g 75 -f 10e6 --freq-bands 12 --start-freq 10e6 --stop-freq 6e9 --duration 2.0 --auto --lo-export True,False,False,False --lo-source internal,companion,external,external
  9. At each frequency step, tune Signal Generator to the displayed frequency + 1 MHz and increase output power by 3 dB.
  10. Analyze terminal output. The "run avg" across all runs should not deviate more than 1 degree and the "stddev" for any run should not deviate more than 1 degree.
  1. Set different IP addresses on each X3x0 and make sure correct FPGA image is loaded on each.
  2. Place first daughterboard in slot A of first X3x0 and second daughterboard in slot A of second X3x0.
  3. Connect host to both X3x0s.
  4. Connect 10 MHz and PPS from Octoclock-G to both X3x0s.
  5. Connect Signal Generator to input of splitter and outputs of the splitter to the RX2 port on each daughterboard.
  6. Set Signal Generator output power at -35 dBm for UBX and -29 dBm for SBX.
  7. From the top of the UHD source, run the command:
    • ./tools/usrptest/uhd_phase_alignment.py --args "addr0=<first X3x0 IP addr>,addr1=<second X3x0 IP addr>,dboard_clock_rate=20e6" --clock-source external --time-source external --sync pps --spec "A:0" --channels 0,1 -s 10e6 -g \<33 if UBX 37.5 if SBX\> -f \<lowest DB freq\> --freq-bands \<# frequency bands\> --start-freq \<lowest freq\> --stop-freq \<highest freq\> --duration 2.0 --auto
  8. At each frequency step, tune Signal Generator to the displayed frequency + 1 MHz and increase output power by 1dB.
  9. Analyze terminal output. The "run avg" across all runs should not deviate more than 2 degrees and the "stddev" for any run should not deviate more than 2 degrees.
N2x0 MIMO with SBX
  1. Set different IP addresses on each N2x0 and make sure correct FPGA image and firmware are loaded.
  2. Connect MIMO cable between devices.
  3. Connect host to master device via 1 GbE.
  4. Connect 10 MHz and PPS from Octoclock-G to master device only.
  6. Set Signal Generator output power at -36 dBm.
    • ./tools/usrptest/usrp_phase_alignment.py --args "addr0=<IP address of master>,addr1=<IP address of slave>" --clock-source external,mimo --time-source external,mimo --sync default --channels 0,1 -s 10e6 -f 400e6 -g 31.5 --freq-bands 7 --start-freq 400e6 --stop-freq 4400e6 --duration 2.0 --auto
  8. At each frequency step, tune Signal Generator to the displayed frequency + 1 MHz and increase output power by 2 dB.
  9. Analyze terminal output. The "run avg" across all runs should not deviate more than 2 degrees and the "stddev" for any run should not deviate more than 5 degrees.
tbd
BISTs
BIST-N310-v1 1xN310 DB-15 GPIO Loopback N300/N310 Manual Procedure N300/N310 Automatic Procedure
BIST-N300-v1 1xN300 DB-15 GPIO Loopback N300/N310 Manual Procedure N300/N310 Automatic Procedure
BIST-E320-v1 1xE320 Type C HDMI Cable + Breakout board with GPIO Loopback E320 Manual Procedure E320 Automatic Procedure
Some of our devices have built-in self-tests (BISTs).
Note: The N300 and N310 have identical BISTs.
  1. Connect the front-panel GPIO loopback to the front panel (see DB15 GPIO Loopback)
  2. Execute the following commands:
     $ n3xx_bist standard # Note: This will run multiple tests
     $ n3xx_bist gpio
  3. Load the AA image from a host computer. Adapt the following command to your system:
     $ uhd_image_loader --args type=n3xx,addr=ni-n3xx-$SERIAL --fpga-path=/path/to/usrp_n310_fpga_AA.bit
  4. The final BIST is the SFP test, and there are three valid ways of executing them. For the purpose of running the test, only one of these needs to be run, and can be chosen based on the available peripherals. a. Connect an SFP loopback module to both SFP0 and SFP1. Run the command n3xx_bist sfp0_loopback sfp1_loopback. b. If only one SFP loopback module is available, connect the loopback module to SFP0 and run n3xx_bist sfp0_loopback. Then, connect the loopback module to SFP1 and run the command n3xx_bist sfp1_loopback. c. If no SFP loopback module is available, connect an SFP cable to both SFP0 and SFP1, and run the command n3xx_bist sfp_loopback. Note while this option is a legitimate substitute to the previous two options, it is of limited use when trying to debug actual SFP issues.
Every test will produce a JSON-serialized dictionary. All tests have passed if the "status" key is "true", or the return code for n3xx_bist is 0.
Note: Keep in mind that after the test, an Aurora image is loaded. If this is not desired, re-run uhd_image_loader to load whatever image is requested.
Assuming the peripherals described in N300/N310 Manual Procedure are all plugged in, the test can trivially be executed automatically by running
    $ n3xx_bist gpio
    $ n3xx_bist sfp_loopback # Or sfp0_loopback and sfp1_loopback
and making sure that all return values are 0.
Required Peripherals
DB15 GPIO Loopback
This is a cable or breakout board which connects to the DB15 connector and loops back the following pins:
  • 0<->6
  • 1<->7
  • 2<->8
  • 3<->9
  • 4<->10
  • 5<->11
  1. Connect the front-panel GPIO loopback to the front panel (see GPIO Loopback)
     $ e320_bist standard # Note: This will run multiple tests
     $ e320_bist gpio
  3. The final BIST is the SFP test. Connect an SFP loopback module to the SFP port. Run the command:
     $ e320_bist sfp_loopback
Every test will produce a JSON-serialized dictionary. All tests have passed if the "status" key is "true", or the return code for e320_bist is 0.
Assuming the peripherals described in E320 Manual Procedure are all plugged in, the test can trivially be executed automatically by running
    $ e320_bist gpio
GPIO Loopback
  1. Type C to Type A HDMI Cable
  2. Breakout Board which connects to the connector and loops back the following pins:
  • 0<->4
  • 1<->5
  • 2<->6
  • 3<->7
Defining R&D Tests
Tests can be added any time to define procedures for pass/fail validation. Any test must include the following:
  • An unambiguous test code. This code consists of three characters that identify the test, a short description of the devices required, and a version suffix. Example: GPS-X310-OCXO-v1 is a GPS-related test, requires an X310 and an OCXO to run, and is version 1 of this test.
  • A manual testing procedure. This must unambiguously define a set of tasks, and clearly identify whether or not a test has failed or passed. Tests do not require any other defined outcome other than 'pass' and 'fail'.
  • Optional, but highly recommended: An automatic test procedure. This must consist of a command, or a script, or a set of commands that can be automatically executed, and that will report a failure condition by means of returning a non-zero return value.
Basic understanding of the operation of USRPs by the test operator should be assumed when authoring test procedures. The descriptions should be as short as possible to fully describe, unambiguously, how to reach a pass/fail conclusion.
Test procedures may be updated at any time. If this happens, a new test code must be generated, with the version number increased. Old test codes are considered deprecated (if there exists a version 2 of a test, version 1 should not be run any more).
